Hello guys, I have an HD11XE with the duplicated image only on the endocavitary transducer, can someone help me? does the resintallation of the software solve?

 

https://1drv.ms/u/s!ApL6sjU_ATpWk_VsiFKdeEX5W9c-dA?e=QF0edc


Quote
admin
(@admin)
Admin Admin
Joined: 5 years ago
Posts: 63
18/04/2021 8:33 am  

Its a probe problem.it happen in convex probes


ReplyQuote
admin
(@admin)
Admin Admin
Joined: 5 years ago
Posts: 63
18/04/2021 8:49 am  

most probably it is a probe issue. In some cases power supply also will make similar issue . 
test with other probe to narrow it down

 

And some time the c5-2 probe make mirror and affect on other probe

 

Yes most it probes in low cases it power supply issue
But you can get another probes to test


ReplyQuote
Douglas Cardoso
(@douglas-cardoso)
Premium Member
Joined: 3 years ago
Posts: 6
12/05/2021 2:14 pm  

replaces the probe and it worked correctly, thank you very much.


ReplyQuote
Douglas Cardoso
(@douglas-cardoso)
Premium Member
Joined: 3 years ago
Posts: 6
19/08/2021 12:54 am  

Hi guys, I managed to repair the transducer in a specialized company and it worked correctly, thank you for your attention.